Knowledge Base Article

Why does the PHY Lite for Parallel Interfaces IP Parameter Editor GUI not produce any error message when input or bidirectional pins are placed in the same lane as RZQ using POD I/O standards?

Description

Due to a problem in Quartus® Prime Pro Edition Software version 24.3, you might see that the PHY Lite for Parallel Interfaces IP Parameter Editor GUI does not produce an error message when input or bidirectional pins are placed in the same lane as the RZQ pin, and the I/O standard is either 1.1-V POD or 1.2-V POD.

When lane 3 is used (RZQ pin index 38), the Quartus® compilation fails with the following error message:

Error(12406): Port I of I/O buffer "phylite_ph2_0_example_design|phylite_ph2_0_example_design|core|arch_inst|phylite_iobufs_inst|ibuf_rzq.ibuf_inst_rzq" must be connected to a top-level pin

However, when lane 5 is used (RZQ pin index 62), the Quartus® compilation passes when it should fail.

 

Resolution

To work around this problem, use a different lane than the RZQ lane. No input or bidirectional pins can be placed in the same lane as RZQ, i.e., lane 3 (pin index 38) or lane 5 (pin index 62), for all supported I/O standards. Only output pins can be placed in the same lane as the RZQ pin.

The problem has been fixed starting with Quartus® Prime Pro Edition software version 25.1.